From vidyagv at gmail.com Thu Sep 1 11:46:49 2005 From: vidyagv at gmail.com (Vidya V) Date: Thu, 1 Sep 2005 11:46:49 -0700 Subject: Request for processinf BCB Message-ID: <616a357905090111462faab54c@mail.gmail.com> Dear Specmat members, I would like to make a request to process 4022-35 PhotoBCB in the following manner: 1. Clean wafer 2. Spin coat adhesion promoter, AP3000 3. Spincoat BCB 4. Softbake for 60 seconds at 65 degrees C 5. Expose 6. Pre-develop bake the wafer for 5 minutes at 60 degrees 7. Develop in the DS3000 developer which is at 40 degrees for 1 minute 8. rinse in room temperature developer for a few minutes 9. Spin dry. The point of concern here could be that , the developer. which is benzene based, needs to be at an elevated temperature for optimum results. PLease let me know how/where this can be done. APpreciate you r co-operation, the MSDS for DS3000 and the processing procedures are attached with this email. Regards, Vidya -------------- next part -------------- An HTML attachment was scrubbed... URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: MSDS_BCB_Developer.pdf Type: application/pdf Size: 25532 bytes Desc: not available URL: -------------- next part -------------- A non-text attachment was scrubbed... Name: Develop_procedure.pdf Type: application/pdf Size: 318866 bytes Desc: not available URL: From edmyers at stanford.edu Fri Sep 2 11:26:37 2005 From: edmyers at stanford.edu (Ed Myers) Date: Fri, 02 Sep 2005 11:26:37 -0700 Subject: nickel silicide In-Reply-To: References: Message-ID: <6.2.1.2.2.20050902111850.0423f7a8@edmyers.pobox.stanford.edu> Ritesh, Ahmet and Venatagirish, SpecMat has reviewed your request for Ni processing. Since you have withdrawn your request for poly deposition on Ni silicide we can approve the remainder of your request. Nickel has been and will remain under the SemicleanB material classification. Regards, At 03:00 PM 8/2/2005, Ritesh J wrote: >Hello > >Currently we 3 labmates from UCLA (riteshj, nvgirish, tura) use nickel >silicided wafers (deposition of nickel and then silicide in AG4108) for >the follwoing processes: >1) After nickel silicide, deposit LTO (~5000A) >2) Etch LTO (contact etch, stop at nickel silicide) in AMT etcher >3) Etch Nitride ((stop at nickel silicide) in AMT etcher >4) Deposit Poly (nickel silicide covered with 5000A LTO (right after >LTO)) (?) >5) Poly etch in P5000 after the previous step (silicide is covered) >6) Deposit Aluminum in Gryphon > >The first three processes had been previously approved for a labmate who had >worked with nickel silicide. If possible, Can you add Nickel to the >standard metal list ? THank you, > >riteshj , tura , nvgirish >(UCLA) From mahnaz at snf.stanford.edu Tue Sep 6 08:27:07 2005 From: mahnaz at snf.stanford.edu (Mahnaz Mansourpour) Date: Tue, 06 Sep 2005 08:27:07 -0700 Subject: [Fwd: approved materials for headway/karlsuss] Message-ID: <431DB54B.3050009@snf.stanford.edu> -------- Original Message -------- Subject: approved materials for headway/karlsuss Date: Fri, 2 Sep 2005 09:43:06 -0700 From: Simon Fishel To: 'Mahnaz Mansourpour' Mahnaz, I was wondering if calcium fluoride glass is an approved substrate for use in the headway and the karlsuss. Please let me know. Thanks, Simon -------------- next part -------------- An HTML attachment was scrubbed... URL: From rissman at stanford.edu Tue Sep 6 08:59:24 2005 From: rissman at stanford.edu (Paul Rissman) Date: Tue, 06 Sep 2005 08:59:24 -0700 Subject: [Fwd: approved materials for headway/karlsuss] Message-ID: <6.2.1.2.2.20050906085836.02bcf938@rissman.pobox.stanford.edu> It is my understanding that other than being mildly hygroscopic, calcium fluoride is relatively intert. >X-Sieve: CMU Sieve 2.2 >Delivered-To: rissman at snf.stanford.edu >Mailing-List: contact specmat-help at snf.stanford.edu; run by ezmlm >X-No-Archive: yes >List-Post: >List-Help: >List-Unsubscribe: >List-Subscribe: >Delivered-To: mailing list specmat at snf.stanford.edu >Date: Tue, 06 Sep 2005 08:27:07 -0700 >From: Mahnaz Mansourpour >Organization: SNF >User-Agent: Mozilla/5.0 (Windows; U; Windows NT 5.0; en-US; rv:1.7.2) >Gecko/20040804 Netscape/7.2 (ax) >X-Accept-Language: en-us, en >To: specmat >Subject: [Fwd: approved materials for headway/karlsuss] > > > >-------- Original Message -------- >Subject: approved materials for headway/karlsuss >Date: Fri, 2 Sep 2005 09:43:06 -0700 >From: Simon Fishel >To: 'Mahnaz Mansourpour' > > > > >Mahnaz, > >I was wondering if calcium fluoride glass is an approved substrate for >use in the headway and the karlsuss. Please let me know. > >Thanks, > >Simon > From mtang at snf.stanford.edu Tue Sep 6 09:12:51 2005 From: mtang at snf.stanford.edu (Mary Tang) Date: Tue, 06 Sep 2005 09:12:51 -0700 Subject: [Fwd: approved materials for headway/karlsuss] In-Reply-To: <431DB54B.3050009@snf.stanford.edu> References: <431DB54B.3050009@snf.stanford.edu> Message-ID: <431DC003.6010907@snf.stanford.edu> Hi Simon/Mahnaz -- Although in principle, it should be safe to use, could we get a little more information? CaF has been used in the lab, but there are a couple of concerns with patterning it... 1. Could we get a process flow? Will the CaF be cleaned in the lab prior to lithography? After lithography, will it be etched in the lab? 2. What are the dimensions of the CaF substrate? (The contact aligners [and even the headway] can tolerate only a limited range of substrate thicknesses.) 3. And just out of curiosity, what are the general dimensions of the features to be patterned and with what alignment precision, if any? Thanks, Mary Mahnaz Mansourpour wrote: > > > -------- Original Message -------- > Subject: approved materials for headway/karlsuss > Date: Fri, 2 Sep 2005 09:43:06 -0700 > From: Simon Fishel > To: 'Mahnaz Mansourpour' > > > >Mahnaz, > >I was wondering if calcium fluoride glass is an approved substrate for >use in the headway and the karlsuss. Please let me know. > >Thanks, > >Simon > > > -- Mary X. Tang, Ph.D. Stanford Nanofabrication Facility CIS Room 136, Mail Code 4070 Stanford, CA 94305 (650)723-9980 mtang at stanford.edu http://snf.stanford.edu From mtang at snf.stanford.edu Tue Sep 6 13:51:34 2005 From: mtang at snf.stanford.edu (Mary Tang) Date: Tue, 06 Sep 2005 13:51:34 -0700 Subject: [Fwd: Re: [Fwd: approved materials for headway/karlsuss]] Message-ID: <431E0156.8030103@snf.stanford.edu> Hi everyone -- Sounds good to me (the thickness should be fine and as long as there is no etching, it should be OK.) Any objections? Mary -- Mary X. Tang, Ph.D. Stanford Nanofabrication Facility CIS Room 136, Mail Code 4070 Stanford, CA 94305 (650)723-9980 mtang at stanford.edu http://snf.stanford.edu -------------- next part -------------- An embedded message was scrubbed... From: "Simon Fishel" Subject: RE: [Fwd: approved materials for headway/karlsuss] Date: Tue, 6 Sep 2005 10:22:23 -0700 Size: 3226 URL: From rissman at stanford.edu Tue Sep 6 14:02:29 2005 From: rissman at stanford.edu (Paul Rissman) Date: Tue, 06 Sep 2005 14:02:29 -0700 Subject: [Fwd: Re: [Fwd: approved materials for headway/karlsuss]] In-Reply-To: <431E0156.8030103@snf.stanford.edu> References: <431E0156.8030103@snf.stanford.edu> Message-ID: <6.2.1.2.2.20050906140128.02e84608@rissman.pobox.stanford.edu> I agree. At 01:51 PM 9/6/2005, Mary Tang wrote: >Hi everyone -- > >Sounds good to me (the thickness should be fine and as long as there is no >etching, it should be OK.) Any objections? > >Mary > >-- >Mary X. Tang, Ph.D. >Stanford Nanofabrication Facility >CIS Room 136, Mail Code 4070 >Stanford, CA 94305 >(650)723-9980 >mtang at stanford.edu >http://snf.stanford.edu > > > >Return-Path: >Delivered-To: mtang at snf.stanford.edu >Received: (qmail 7240 invoked from network); 6 Sep 2005 17:22:35 -0000 >Received: from smtp1.stanford.edu (171.67.16.123) > by snf.stanford.edu with SMTP; 6 Sep 2005 17:22:35 -0000 >Received: from your495cd88010 (DNab407ae9.Stanford.EDU [171.64.122.233]) > by smtp1.Stanford.EDU (8.12.11/8.12.11) with ESMTP id j86HMQm3029036 > for ; Tue, 6 Sep 2005 10:22:33 -0700 >From: "Simon Fishel" >To: "'Mary Tang'" >Subject: RE: [Fwd: approved materials for headway/karlsuss] >Date: Tue, 6 Sep 2005 10:22:23 -0700 >Message-ID: <000a01c5b307$8c158250$e97a40ab at your495cd88010> >MIME-Version: 1.0 >Content-Type: text/plain; > charset="us-ascii" >Content-Transfer-Encoding: 7bit >X-Priority: 3 (Normal) >X-MSMail-Priority: Normal >X-Mailer: Microsoft Outlook, Build 10.0.2627 >X-MimeOLE: Produced By Microsoft MimeOLE V6.00.2900.2527 >In-Reply-To: <431DC003.6010907 at snf.stanford.edu> >Importance: Normal > >Mary, > >The samples are cleaned in our lab prior to bringing them into SNF, so >we will just bake them in the Singe oven before coating them in the >Headway. Depending on the resist we end up using, most likely SPR220-7, >we will then do the appropriate pre-exposure bake, expose the samples on >the Karlsuss, and develop them in the solvent bench. The resist itself >serves as our features, so there is no etch involved. > >The samples are circular, 3/4 inch in diameter, and about 0.5 mm thick. >We are only patterning one layer, and our features are on the order of >millimeters, so very little alignment precision is needed. > >Let me know if you need any more information. > >Simon > >-----Original Message----- >From: Mary Tang [mailto:mtang at snf.stanford.edu] >Sent: Tuesday, September 06, 2005 9:13 AM >To: Mahnaz Mansourpour >Cc: specmat; sfishel at stanford.edu >Subject: Re: [Fwd: approved materials for headway/karlsuss] > >Hi Simon/Mahnaz -- > >Although in principle, it should be safe to use, could we get a little >more information? CaF has been used in the lab, but there are a couple >of concerns with patterning it... > >1. Could we get a process flow? Will the CaF be cleaned in the lab >prior to lithography? After lithography, will it be etched in the lab? >2. What are the dimensions of the CaF substrate? (The contact aligners > >[and even the headway] can tolerate only a limited range of substrate >thicknesses.) >3. And just out of curiosity, what are the general dimensions of the >features to be patterned and with what alignment precision, if any? > >Thanks, > >Mary > >Mahnaz Mansourpour wrote: > > > > > > > -------- Original Message -------- > > Subject: approved materials for headway/karlsuss > > Date: Fri, 2 Sep 2005 09:43:06 -0700 > > From: Simon Fishel > > To: 'Mahnaz Mansourpour' > > > > > > > >Mahnaz, > > > >I was wondering if calcium fluoride glass is an approved substrate for > >use in the headway and the karlsuss. Please let me know. > > > >Thanks, > > > >Simon > > > > > > > > >-- >Mary X. Tang, Ph.D. >Stanford Nanofabrication Facility >CIS Room 136, Mail Code 4070 >Stanford, CA 94305 >(650)723-9980 >mtang at stanford.edu >http://snf.stanford.edu From mtang at snf.stanford.edu Tue Sep 6 14:26:14 2005 From: mtang at snf.stanford.edu (Mary Tang) Date: Tue, 06 Sep 2005 14:26:14 -0700 Subject: [Fwd: Re: [Fwd: approved materials for headway/karlsuss]] In-Reply-To: <6.2.1.2.2.20050906140128.02e84608@rissman.pobox.stanford.edu> References: <431E0156.8030103@snf.stanford.edu> <6.2.1.2.2.20050906140128.02e84608@rissman.pobox.stanford.edu> Message-ID: <431E0976.6020800@snf.stanford.edu> Hi Simon -- Three of us agree (Paul, Mahnaz, and me.) I think that's a SpecMat quorum. Your request is filed and approved. Thanks! Mary Paul Rissman wrote: > I agree. > > At 01:51 PM 9/6/2005, Mary Tang wrote: > >> Hi everyone -- >> >> Sounds good to me (the thickness should be fine and as long as there >> is no etching, it should be OK.) Any objections? >> >> Mary >> >> -- >> Mary X. Tang, Ph.D. >> Stanford Nanofabrication Facility >> CIS Room 136, Mail Code 4070 >> Stanford, CA 94305 >> (650)723-9980 >> mtang at stanford.edu >> http://snf.stanford.edu >> >> >> >> Return-Path: >> Delivered-To: mtang at snf.stanford.edu >> Received: (qmail 7240 invoked from network); 6 Sep 2005 17:22:35 -0000 >> Received: from smtp1.stanford.edu (171.67.16.123) >> by snf.stanford.edu with SMTP; 6 Sep 2005 17:22:35 -0000 >> Received: from your495cd88010 (DNab407ae9.Stanford.EDU [171.64.122.233]) >> by smtp1.Stanford.EDU (8.12.11/8.12.11) with ESMTP id >> j86HMQm3029036 >> for ; Tue, 6 Sep 2005 10:22:33 -0700 >> From: "Simon Fishel" >> To: "'Mary Tang'" >> Subject: RE: [Fwd: approved materials for headway/karlsuss] >> Date: Tue, 6 Sep 2005 10:22:23 -0700 >> Message-ID: <000a01c5b307$8c158250$e97a40ab at your495cd88010> >> MIME-Version: 1.0 >> Content-Type: text/plain; >> charset="us-ascii" >> Content-Transfer-Encoding: 7bit >> X-Priority: 3 (Normal) >> X-MSMail-Priority: Normal >> X-Mailer: Microsoft Outlook, Build 10.0.2627 >> X-MimeOLE: Produced By Microsoft MimeOLE V6.00.2900.2527 >> In-Reply-To: <431DC003.6010907 at snf.stanford.edu> >> Importance: Normal >> >> Mary, >> >> The samples are cleaned in our lab prior to bringing them into SNF, so >> we will just bake them in the Singe oven before coating them in the >> Headway. Depending on the resist we end up using, most likely SPR220-7, >> we will then do the appropriate pre-exposure bake, expose the samples on >> the Karlsuss, and develop them in the solvent bench. The resist itself >> serves as our features, so there is no etch involved. >> >> The samples are circular, 3/4 inch in diameter, and about 0.5 mm thick. >> We are only patterning one layer, and our features are on the order of >> millimeters, so very little alignment precision is needed. >> >> Let me know if you need any more information. >> >> Simon >> >> -----Original Message----- >> From: Mary Tang [mailto:mtang at snf.stanford.edu] >> Sent: Tuesday, September 06, 2005 9:13 AM >> To: Mahnaz Mansourpour >> Cc: specmat; sfishel at stanford.edu >> Subject: Re: [Fwd: approved materials for headway/karlsuss] >> >> Hi Simon/Mahnaz -- >> >> Although in principle, it should be safe to use, could we get a little >> more information? CaF has been used in the lab, but there are a couple >> of concerns with patterning it... >> >> 1. Could we get a process flow? Will the CaF be cleaned in the lab >> prior to lithography? After lithography, will it be etched in the lab? >> 2. What are the dimensions of the CaF substrate? (The contact aligners >> >> [and even the headway] can tolerate only a limited range of substrate >> thicknesses.) >> 3. And just out of curiosity, what are the general dimensions of the >> features to be patterned and with what alignment precision, if any? >> >> Thanks, >> >> Mary >> >> Mahnaz Mansourpour wrote: >> >> > >> > >> > -------- Original Message -------- >> > Subject: approved materials for headway/karlsuss >> > Date: Fri, 2 Sep 2005 09:43:06 -0700 >> > From: Simon Fishel >> > To: 'Mahnaz Mansourpour' >> > >> > >> > >> >Mahnaz, >> > >> >I was wondering if calcium fluoride glass is an approved substrate for >> >use in the headway and the karlsuss. Please let me know. >> > >> >Thanks, >> > >> >Simon >> > >> > >> > >> >> >> -- >> Mary X. Tang, Ph.D. >> Stanford Nanofabrication Facility >> CIS Room 136, Mail Code 4070 >> Stanford, CA 94305 >> (650)723-9980 >> mtang at stanford.edu >> http://snf.stanford.edu > -- Mary X. Tang, Ph.D. Stanford Nanofabrication Facility CIS Room 136, Mail Code 4070 Stanford, CA 94305 (650)723-9980 mtang at stanford.edu http://snf.stanford.edu From edmyers at stanford.edu Tue Sep 6 14:29:41 2005 From: edmyers at stanford.edu (Ed Myers) Date: Tue, 06 Sep 2005 14:29:41 -0700 Subject: Fwd: Selenious acid in fab Message-ID: <6.2.1.2.2.20050906142916.041acd08@edmyers.pobox.stanford.edu> SpecMat Members, Here is another request..... >X-Sieve: CMU Sieve 2.2 >X-Mailer: QUALCOMM Windows Eudora Version 5.2.1 >Date: Tue, 06 Sep 2005 12:44:18 -0700 >To: Ed Myers >From: Carl Faulkner >Subject: Selenious acid in fab > >Ed, > >I would like to try duplicating a result from a paper using selenium to >passivate the silicon surface. This requires dissolving small quantities >of selenious acid in methanol (ending PH 4.5 or ~0.5mg Se/cc of >solution). I would do this in personal labware (and either WBgeneral or >WBsolvent?) and the resulting wafer would only go into the Innotec >afterwards to be capped with Al (at that point there should be less than a >monolayer of Se terminating the Si dangling bonds present). At the >moment, I only wish to try it so the quantities would be small. If it >works, we would wish to run more wafers with different metals, though most >of the metals I can think of using are ones I deposit at Berkeley. The >problem I see is that selenious acid is toxic and would generate hazardous >waste. For the initial experiment, I would only generate ~300ml of >solution plus additional methanol to clean out the beaker. > >MSDS at https://fscimage.fishersci.com/msds/88641.htm > >Carl From lll2002 at stanford.edu Thu Sep 8 10:45:19 2005 From: lll2002 at stanford.edu (Mr Liangliang Li) Date: Thu, 8 Sep 2005 10:45:19 -0700 Subject: new process by using 3 % NaOH Message-ID: <1126201519.432078af1f8d1@webmail.stanford.edu> Dear Sir/Madam: How are you? I want to use 3% NaOH solution to strip the resist off from an 8-inch substrate, I hope I can get your permission. the following is the detailed process I want to do on the wet bench general: 1: I will get my own 3% NaOH solution and transfer it into the cleaning room to the wet bench. I will mix NaOH powder with water or dilute 50% NaOH solution outside of the cleanining room, to get 3% solution. 2: I will get a large beaker (9 or 10 inch round size and 6-8 inch tall) and bring it into the cleaning room. 3: one 8-inch Printed circuit board substrate with a mask (made of negative resist) on the top surface will be shipped from Intel. In our own lab, my magnetic material will be deposited to cover the whole substrate. And then, I will bring the substrate to the wet bench. 4: in the wet bench general, I will soak the 8-inch substrate into the 3% NaOH solution inside the big beaker and heat to 45 degree C for 30 mins. The resist will be stripped off from the substate. 5: After stripping, I will clean the wet bench and waste accordding to the manual, the 8-inch substate will be taked out of the cleaning room immediately and shipped back to Intel. In our own lab, I have tested the stability of the substate in 3% NaOH solution at 45 C. It is really stable. The hot plate size is 8 inch by 8 inch and the tank is 11-12 inch size. So we have enough space. Dear Sir/Madam, if you think this process is ok in the wet bench general, please let me know. Thank you very much best regards liangliang li From mtang at snf.stanford.edu Thu Sep 8 11:02:55 2005 From: mtang at snf.stanford.edu (Mary Tang) Date: Thu, 8 Sep 2005 11:02:55 -0700 (PDT) Subject: new process by using 3 % NaOH In-Reply-To: <1126201519.432078af1f8d1@webmail.stanford.edu> Message-ID: Hi all -- I think this, in general, sounds OK. The conditions that would make me feel better are: 1. Waste be collected locally and properly labeled for disposal (this sounds like a lift-off process, and as we don't know what the materials are, I don't feel comfortable putting it down the AWN.) 2. Mixing should be done outside the lab. We should reiterate the procedures and policies for transporting chemicals to SNF, especially as I'm not sure who this is, much less that he is aware of these policies in practice. How do these sound to you? Mary On Thu, 8 Sep 2005, Mr Liangliang Li wrote: > Dear Sir/Madam: > How are you? > I want to use 3% NaOH solution to strip the resist off from an 8-inch > substrate, I hope I can get your permission. the following is the detailed > process I want to do on the wet bench general: > > 1: I will get my own 3% NaOH solution and transfer it into the cleaning room > to the wet bench. I will mix NaOH powder with water or dilute 50% NaOH solution > outside of the cleanining room, to get 3% solution. > 2: I will get a large beaker (9 or 10 inch round size and 6-8 inch tall) and > bring it into the cleaning room. > 3: one 8-inch Printed circuit board substrate with a mask (made of negative > resist) on the top surface will be shipped from Intel. In our own lab, my > magnetic material will be deposited to cover the whole substrate. And then, I > will bring the substrate to the wet bench. > 4: in the wet bench general, I will soak the 8-inch substrate into the 3% > NaOH solution inside the big beaker and heat to 45 degree C for 30 mins. The > resist will be stripped off from the substate. > 5: After stripping, I will clean the wet bench and waste accordding to the > manual, the 8-inch substate will be taked out of the cleaning room immediately > and shipped back to Intel. > > In our own lab, I have tested the stability of the substate in 3% NaOH > solution at 45 C. It is really stable. The hot plate size is 8 inch by 8 inch > and the tank is 11-12 inch size. So we have enough space. > > Dear Sir/Madam, if you think this process is ok in the wet bench general, > please let me know. > Thank you very much > > best regards > liangliang li > From rissman at stanford.edu Thu Sep 8 11:42:02 2005 From: rissman at stanford.edu (Paul Rissman) Date: Thu, 08 Sep 2005 11:42:02 -0700 Subject: new process by using 3 % NaOH In-Reply-To: References: <1126201519.432078af1f8d1@webmail.stanford.edu> Message-ID: <6.2.1.2.2.20050908114129.02b5c438@rissman.pobox.stanford.edu> I agree with Mary. Don't we need to know what will be etched for disposal reasons? At 11:02 AM 9/8/2005, Mary Tang wrote: >Hi all -- > >I think this, in general, sounds OK. The conditions that would make me >feel better are: > >1. Waste be collected locally and properly labeled for disposal (this >sounds like a lift-off process, and as we don't know what the materials >are, I don't feel comfortable putting it down the AWN.) > >2. Mixing should be done outside the lab. We should reiterate the >procedures and policies for transporting chemicals to SNF, especially as >I'm not sure who this is, much less that he is aware of these policies in >practice. > >How do these sound to you? > >Mary > >On Thu, 8 Sep 2005, Mr Liangliang Li wrote: > > > Dear Sir/Madam: > > How are you? > > I want to use 3% NaOH solution to strip the resist off from an 8-inch > > substrate, I hope I can get your permission. the following is the detailed > > process I want to do on the wet bench general: > > > > 1: I will get my own 3% NaOH solution and transfer it into the > cleaning room > > to the wet bench. I will mix NaOH powder with water or dilute 50% NaOH > solution > > outside of the cleanining room, to get 3% solution. > > 2: I will get a large beaker (9 or 10 inch round size and 6-8 inch > tall) and > > bring it into the cleaning room. > > 3: one 8-inch Printed circuit board substrate with a mask (made of > negative > > resist) on the top surface will be shipped from Intel. In our own lab, my > > magnetic material will be deposited to cover the whole substrate. And > then, I > > will bring the substrate to the wet bench. > > 4: in the wet bench general, I will soak the 8-inch substrate into > the 3% > > NaOH solution inside the big beaker and heat to 45 degree C for 30 > mins. The > > resist will be stripped off from the substate. > > 5: After stripping, I will clean the wet bench and waste accordding > to the > > manual, the 8-inch substate will be taked out of the cleaning room > immediately > > and shipped back to Intel. > > > > In our own lab, I have tested the stability of the substate in 3% NaOH > > solution at 45 C. It is really stable. The hot plate size is 8 inch by > 8 inch > > and the tank is 11-12 inch size. So we have enough space. > > > > Dear Sir/Madam, if you think this process is ok in the wet bench > general, > > please let me know. > > Thank you very much > > > > best regards > > liangliang li > > From mtang at snf.stanford.edu Thu Sep 8 12:09:55 2005 From: mtang at snf.stanford.edu (Mary Tang) Date: Thu, 8 Sep 2005 12:09:55 -0700 (PDT) Subject: new process by using 3 % NaOH In-Reply-To: <6.2.1.2.2.20050908114129.02b5c438@rissman.pobox.stanford.edu> Message-ID: Hi Paul -- Good question. The hazardous waste should be labelled appropriately by the user. Since he says he's done it before and it's stable, I'm perhaps presuming that he knows what he's doing. But you're right -- we should ask because it's possible that the waste may require special attention (like it might explode if it isn't addressed in time -- this actually happened to some people in the lab above mine when I was in grad school -- blew out a bunch of windows and trashed the chemical bench, though luckily no one was seriously injured.) We can ask him. M On Thu, 8 Sep 2005, Paul Rissman wrote: > I agree with Mary. Don't we need to know what will be etched for disposal > reasons? > > At 11:02 AM 9/8/2005, Mary Tang wrote: > >Hi all -- > > > >I think this, in general, sounds OK. The conditions that would make me > >feel better are: > > > >1. Waste be collected locally and properly labeled for disposal (this > >sounds like a lift-off process, and as we don't know what the materials > >are, I don't feel comfortable putting it down the AWN.) > > > >2. Mixing should be done outside the lab. We should reiterate the > >procedures and policies for transporting chemicals to SNF, especially as > >I'm not sure who this is, much less that he is aware of these policies in > >practice. > > > >How do these sound to you? > > > >Mary > > > >On Thu, 8 Sep 2005, Mr Liangliang Li wrote: > > > > > Dear Sir/Madam: > > > How are you? > > > I want to use 3% NaOH solution to strip the resist off from an 8-inch > > > substrate, I hope I can get your permission. the following is the detailed > > > process I want to do on the wet bench general: > > > > > > 1: I will get my own 3% NaOH solution and transfer it into the > > cleaning room > > > to the wet bench. I will mix NaOH powder with water or dilute 50% NaOH > > solution > > > outside of the cleanining room, to get 3% solution. > > > 2: I will get a large beaker (9 or 10 inch round size and 6-8 inch > > tall) and > > > bring it into the cleaning room. > > > 3: one 8-inch Printed circuit board substrate with a mask (made of > > negative > > > resist) on the top surface will be shipped from Intel. In our own lab, my > > > magnetic material will be deposited to cover the whole substrate. And > > then, I > > > will bring the substrate to the wet bench. > > > 4: in the wet bench general, I will soak the 8-inch substrate into > > the 3% > > > NaOH solution inside the big beaker and heat to 45 degree C for 30 > > mins. The > > > resist will be stripped off from the substate. > > > 5: After stripping, I will clean the wet bench and waste accordding > > to the > > > manual, the 8-inch substate will be taked out of the cleaning room > > immediately > > > and shipped back to Intel. > > > > > > In our own lab, I have tested the stability of the substate in 3% NaOH > > > solution at 45 C. It is really stable. The hot plate size is 8 inch by > > 8 inch > > > and the tank is 11-12 inch size. So we have enough space. > > > > > > Dear Sir/Madam, if you think this process is ok in the wet bench > > general, > > > please let me know. > > > Thank you very much > > > > > > best regards > > > liangliang li > > > > From edmyers at stanford.edu Fri Sep 9 11:46:11 2005 From: edmyers at stanford.edu (Ed Myers) Date: Fri, 09 Sep 2005 11:46:11 -0700 Subject: new process by using 3 % NaOH In-Reply-To: References: <6.2.1.2.2.20050908114129.02b5c438@rissman.pobox.stanford.edu> Message-ID: <6.2.1.2.2.20050909114141.043fc150@edmyers.pobox.stanford.edu> All, The user stopped by and is anxious to get approval. I feel we can approve this request under the following guidelines. 1) Use his own, dedicated beakers. 2) All mixing of powders will be done outside of the lab. 3) Local collection and labeling of the waste. The film he will be lifting off is a magnetic layer, but I do not know the composition. If I don't hear any why we should not approve this request by mid-afternoon today (9/9), I will send him an approval email. Ed At 12:09 PM 9/8/2005, you wrote: >Hi Paul -- > >Good question. The hazardous waste should be labelled appropriately by >the user. Since he says he's done it before and it's stable, I'm perhaps >presuming that he knows what he's doing. But you're right -- we should >ask because it's possible that the waste may require special attention >(like it might explode if it isn't addressed in time -- this actually >happened to some people in the lab above mine when I was in grad school >-- blew out a bunch of windows and trashed the chemical bench, though >luckily no one was seriously injured.) We can ask him. > >M > >On Thu, 8 Sep 2005, Paul Rissman wrote: > > > I agree with Mary. Don't we need to know what will be etched for disposal > > reasons? > > > > At 11:02 AM 9/8/2005, Mary Tang wrote: > > >Hi all -- > > > > > >I think this, in general, sounds OK. The conditions that would make me > > >feel better are: > > > > > >1. Waste be collected locally and properly labeled for disposal (this > > >sounds like a lift-off process, and as we don't know what the materials > > >are, I don't feel comfortable putting it down the AWN.) > > > > > >2. Mixing should be done outside the lab. We should reiterate the > > >procedures and policies for transporting chemicals to SNF, especially as > > >I'm not sure who this is, much less that he is aware of these policies in > > >practice. > > > > > >How do these sound to you? > > > > > >Mary > > > > > >On Thu, 8 Sep 2005, Mr Liangliang Li wrote: > > > > > > > Dear Sir/Madam: > > > > How are you? > > > > I want to use 3% NaOH solution to strip the resist off from an > 8-inch > > > > substrate, I hope I can get your permission. the following is the > detailed > > > > process I want to do on the wet bench general: > > > > > > > > 1: I will get my own 3% NaOH solution and transfer it into the > > > cleaning room > > > > to the wet bench. I will mix NaOH powder with water or dilute 50% NaOH > > > solution > > > > outside of the cleanining room, to get 3% solution. > > > > 2: I will get a large beaker (9 or 10 inch round size and 6-8 inch > > > tall) and > > > > bring it into the cleaning room. > > > > 3: one 8-inch Printed circuit board substrate with a mask (made of > > > negative > > > > resist) on the top surface will be shipped from Intel. In our own > lab, my > > > > magnetic material will be deposited to cover the whole substrate. And > > > then, I > > > > will bring the substrate to the wet bench. > > > > 4: in the wet bench general, I will soak the 8-inch substrate into > > > the 3% > > > > NaOH solution inside the big beaker and heat to 45 degree C for 30 > > > mins. The > > > > resist will be stripped off from the substate. > > > > 5: After stripping, I will clean the wet bench and waste accordding > > > to the > > > > manual, the 8-inch substate will be taked out of the cleaning room > > > immediately > > > > and shipped back to Intel. > > > > > > > > In our own lab, I have tested the stability of the substate in > 3% NaOH > > > > solution at 45 C. It is really stable. The hot plate size is 8 inch by > > > 8 inch > > > > and the tank is 11-12 inch size. So we have enough space. > > > > > > > > Dear Sir/Madam, if you think this process is ok in the wet bench > > > general, > > > > please let me know. > > > > Thank you very much > > > > > > > > best regards > > > > liangliang li > > > > > > From edmyers at stanford.edu Mon Sep 12 13:32:44 2005 From: edmyers at stanford.edu (Ed Myers) Date: Mon, 12 Sep 2005 13:32:44 -0700 Subject: SpecMat Logsheet, 9/13/05 Message-ID: <6.2.1.2.2.20050912133100.04481cd8@edmyers.pobox.stanford.edu> All, Attached is the SpecMat logsheet outlining all activities since the last meeting. We only have a few items, so Tuesday's meeting should be fairly quick. Ed -------------- next part -------------- A non-text attachment was scrubbed... Name: SpecMat Logsheet.xls Type: application/octet-stream Size: 104960 bytes Desc: not available URL: From ForEducation at academic-software-outlet.com Sun Sep 18 13:36:06 2005 From: ForEducation at academic-software-outlet.com (Academic Sales Department) Date: Sun, 18 Sep 2005 13:36:06 -0700 Subject: Deep Software Discounts for Student, Teachers, Staff, Schools Message-ID: <2b307a9f427af3612eee25a4d6eb05@none-r8voc5boyu> Adobe Acrobat 7.0 Professional at 69% OFF, Macromedia Dreamvweaver MX 2004 at 67% OFF, Microsoft Office 2003 Standard at 63% OFF, Adobe Illustrator CS2 12.0 at 76% OFF, EndNote 8.0 at 67% OFF, !!!!!!!!!!!!!BACK-TO-SCHOOL SALE!!!!!!!!!!!!!!!!!! *** EXCLUSIVELY available to Qualified Students, Faculty, and Staff of K-12 and Higher Education institutions. (see below for details) *** COMPUTER PRODUCTS FOR EDUCATION, Inc. is pleased to offer to qualified faculty, staff, and students, the best service and prices on all ACADEMIC EDITION SOFTWARE from MICROSOFT, ADOBE, MACROMEDIA SYMANTEC, COREL, AUTODESK, ENDNOTE, FILEMAKER, and several OTHER MAJOR SOFTWARE MANUFACTURERS - AT DISCOUNTS UP TO 91% OFF STANDARD RETAIL PRICES!!! To order, please visit our website at: www.academic-software-outlet.com or call 800-679-7007. ****** SALE PRICES VALID THROUGH SEPTEMBER 30, 2005 ******** ---------------------- Education Standard You ADOBE - SALE! Price Retail Save! ---------------------- --------- ------ ----- Creative Suite Premium CS2** $379.95 $999 62% Acrobat 7.0 Standard $89.95 $299 78% Acrobat 7.0 Professional* $118.95 $449 76% Illustrator CS2 12.0 $98.95 $499 80% InDesign CS2 4.0 $189.95 $699 72% PageMaker 7.0.2 $289.95 $499 42% Photoshop CS2 9.0 $279.95 $649 55% Photoshop Elements 3.0 $68.95 $99 33% Premiere Pro 1.5 $244.95 $699 65% Video Collection Standard 2.5 $479.95 $999 52% Video Collection Pro 2.6 $869.95 $999 52% * Plus additional $30 Faculty Rebate. **Includes: InDesign 4.0,Photoshop 9.0, Illustrator 12.0, Golive 8.0, Acrobat 7.0 Pro Visit: www.academic-software-outlet.com ---------------------- Education Standard You MACROMEDIA - New Releases! Price Retail Save! ---------------------- --------- ------ ----- Studio 8 $288.95 $999 71% Dreamweaver 8 $198.95 $399 50% Flash 8 $247.95 $699 65% Freehand MX 2004 $ 94.95 $399 75% Director MX 2004 $489.95 $1999 59% Captivate 1.0 $198.95 $499 60% Contribute 3.0 $ 89.95 $149 40% Visit: www.academic-software-outlet.com --------------------------- Education Standard You MICROSOFT - SALE!: Price Retail Save! --------------------------- --------- ------ ----- Office 2003 Professional $178.95 $499 64% Office 2003 Std Students&Teachers $144.95 $399 63% Office Mac 2004 Students&Teachers $148.95 $499 70% FrontPage 2003 $98.95 $199 50% OneNote 2003 $48.95 $199 75% Publisher 2003 $97.95 $169 42% Project 2003 Professional $189.95 $999 80% Project 2003 Standard $69.95 $599 88% Student 2006 $89.95 $ 99 10% Visio 2003 Professional $148.95 $499 70% Visio 2003 Standard $64.95 $199 66% Visual Studio.Net Pro 2003 $92.95 $1079 91% Windows XP Professional Upg $94.95 $199 52% *Office 2003 Professional includes: Word, Excel, PowerPoint, Outlook, Access, and Publisher. **Office 2003 for Students & Teachers includes: Word, Excel, Powerpoint, and Outlook. Visit: www.academic-software-outlet.com --------------------------- Education Standard You AutoDesk Price Retail Save! --------------------------- --------- ------ ----- AutoCAD 2006 Full Version 2D+3D* $379.95 $3750 90% AutoCAD 2006 LT $149.95 $899 83% AutoCAD Architectural Dsktp 2006* $379.95 $4695 92% AutoCAD Building systems 2006 $469.95 $5995 92% AutoCAD Civil 3D 2006* $469.95 $4195 91% AutoCAD Inventor Professional 2006 $469.95 $5195 91% AutoCAD Land Desktop 2006* $469.95 $4995 91% AutoCAD Map 3D 2006* $469.95 $4795 90% AutoCAD Revit Series 8 $469.95 $4995 91% AutoCAD Viz 2006* $289.95 $1995 85% Discreet Combustion 4.0 $289.95 $995 71% Discreet 3ds Max 7 $679.95 $3495 81% *Price for Students, Faculty, & Staff for personal use - pricing for schools for institutional use is different. Please call 800-679-7007 for more information. Visit: www.academic-software-outlet.com --------------------------- Education Standard You AVID Price Retail Save! --------------------------- --------- ------ ----- Avid Xpress Studio Academic $294.95 $1695 83% Avid Xpress Studio HD Essentials $989.95 $3495 72% Avid SoftImage XSI 4.0 $294.95 $1695 83% Avid Mojo Academic $994.95 $1695 41% Visit: www.academic-software-outlet.com --------------------------- Education Standard You PINNACLE: Price Retail Save! --------------------------- --------- ------ ----- Liquid Edition 6 Adv Acad Toolkit $249.95 $999 75% Pinnacle Studio Plus Acad Toolkit $ 79.95 $ 99 19% Visit: www.academic-software-outlet.com --------------------------- Education Standard You SONY: Price Retail Save! --------------------------- --------- ------ ----- Acid Pro 5.0 $149.95 $299 51% Sound Forge 8 w/CD Architect $149.95 $299 50% Vegas 6.0 $199.95 $499 55% Vegas 6.0+DVD Prod Suite $299.95 $674 55% Visit: www.academic-software-outlet.com --------------------------- Education Standard You SYMANTEC: Price Retail Save! --------------------------- --------- ------ ----- Norton Antivirus 2005 $36.95 $49 26% Norton Antivirus 10.0 Mac $54.95 $69 21% Norton Systemworks 2005 $60.95 $69 13% Norton Utilities 8.0 Mac $79.95 $99 20% --------------------------- Education Standard You COREL: Price Retail Save! --------------------------- --------- ------ ----- Corel WordPerfect Office 12 $98.95 $299 67% Corel Draw 12.0 $97.95 $399 72% Corel Painter 8.0 $97.95 $299 67% Visit: www.academic-software-outlet.com --------------------------- Education Standard You EndNote: Price Retail Save! --------------------------- --------- ------ ----- EndNote 9.0-Students $ 99.95 $299 63% EndNote 9.0-Teacher/Schools $199.95 $299 33% Visit: www.academic-software-outlet.com For more information, visit our website at: www.academic-software-outlet.com or call 800-679-7007. PURCHASE ORDERS may be FAXED to: 800-679-6996 (727-530-7994). Educatioal Software has the exact same features and functionality as Commerical Full-Versions of the software. Visit our website to view hundreds of other software products available from CPE at similar discounts. All software products sold by CPE are authentic original software from the manufacturer. THESE ARE NOT PIRATED COPIES. ALL SOFTWARE COMES IN ORIGINAL MANUFACTURER'S BOXES AND INCLUDES A VALID VERIFIABLE LICENSE. ---------- LICENSING: ---------- For school purchases of five to ten (5-10) or more units, depending on the product, please call 800-679-7007 for infomation even deeper discounts on license packs. --------------------------- QUALIFIED EDUCATIONAL BUYERS: --------------------------- The following are defined as Qualified Education Buyers. Qualifed Educational Buyers must provide the below verification upon making any purchase. 1. Students of Higher Education - All enrolled college, junior college, community college, technical school, vocational school, and university students. 2. K-12 Students - K-12 students are eligible for most, but not all, academic-edition software products. 3. Teachers - All K-12 school and Higher Education institution teachers. 4. Faculty - All K-12 school and Higher Education institution faculty. 5. Staff - All K-12 school and Higher Education institution staff. 6. Schools - All elementary, middle and high schools (K-12 schools); vocational and technical schools; correspondence schools, including Internet correspondence schools, and all colleges, including junior and community colleges, and universities. 7. Home Schools - Home schools are now eligible to purchase most Academic Edition software. Home schools must be approved on a case by case basis. Please call for more details. ------------------------- VERIFICATION: ------------------------- Purchasers must provide fax-verification of status as being a current faculty, staff, or student. After placing your order, you simply fax to CPE either: (a) a copy of a current picture School I.D. Card; or, (b) a current paycheck stub with an alternative picture I.D. (drivers license, etc.) Sensitive information may be blacked out; or (c) Schools may purchase by faxing a valid school purchase order. For more details, call us or visit our website. ------------------------- CPE is an Authorized Education Reseller for Microsoft, Adobe, Macromedia, AutoDesk, Corel, Sony, Avid, Pinnacle, Symantec and many other major software manufacturers. CPE is a national software distributor committed to providing the best prices and the best customer service to the Education Community!! Prices generally remain unchanged until the end of the month. However, all prices and availability are subject to change without notice, due to factors outside our control. __________________________ If you would like to receive monthly updates of special offers, and pricing updates, please subscribe to our monthly mailer by clicking on the following link: http://www.academic-software-outlet.com/add.asp?a=add&e=specmat at snf.stanford.edu __________________________ U-N-S-U-S-C-R-1-B-E LINK: We hope you have found this message valuable. However, if you do not wish to recieve any more offers from CPE, please click on the following link: http://www.academic-software-outlet.com/rem.asp?a=remv&e=specmat at snf.stanford.edu Or call 800-679-7007 and ask to be deleted from our list. __________________________ Computer Products for Education 5325 140th Avenue North Clearwater, Florida 33760 Tel: 800-679-7007 / 727-530-1709 Fax: 800-679-6996 / 727-530-7994 ___________________ THANK YOU AND HAVE A NICE DAY!! 42108182178202 From bwchui at yahoo.com Mon Sep 19 12:01:53 2005 From: bwchui at yahoo.com (Benjamin Chui) Date: Mon, 19 Sep 2005 12:01:53 -0700 (PDT) Subject: Aluminum nitride approval request Message-ID: <20050919190153.19510.qmail@web40621.mail.yahoo.com> Hello specmat team, I'd like to request approval to bring wafers with deposited aluminum nitride (AlN) film into the SNF clean room. I hope this is OK since users are specifically using AlN in place of PZT for its CMOS compatibility in building piezoelectric sensors. The company depositing the AlN also cites this advantage. Thanks, Ben Chui From service at creditunion.coop Mon Sep 19 21:12:58 2005 From: service at creditunion.coop (service at creditunion.coop) Date: Mon, 19 Sep 2005 21:12:58 -0700 Subject: Notice Message-ID: An HTML attachment was scrubbed... URL: From edmyers at stanford.edu Tue Sep 20 09:42:32 2005 From: edmyers at stanford.edu (Ed Myers) Date: Tue, 20 Sep 2005 09:42:32 -0700 Subject: Aluminum nitride approval request In-Reply-To: <20050919230600.7230.qmail@web40629.mail.yahoo.com> References: <20050919230600.7230.qmail@web40629.mail.yahoo.com> Message-ID: <6.2.1.2.2.20050920093600.04f73ab0@edmyers.pobox.stanford.edu> Ben, In general the company saying they are CMOS clean is not sufficient. There are facilities which SpecMat has approved as CMOS clean, and wafers from these companies can be processed in our clean equipment group. If the wafers do not come from these companies, we can certify there level of contamination by TXRF and then they will be allowed in the clean equipment set. We did have one large company come in with wafers they said were CMOS clean, but from an unknown vendor. We requested TXRF data and found the wafers to be heavily Cu contaminated. The company had two chambers for the deposition and only one of them showed the high levels. Ultimately the deposition company was forced to do a major clean up on the chamber before they were allowed to process any more wafers. SpecMat will need more information on the origin of the wafers or verification of their level of contamination before they will be approved for the clean equipment set. Ed At 04:06 PM 9/19/2005, you wrote: >Ed, > > I also got confirmation from the company doing the Mo-AlN >deposition that the films are CMOS clean, without gold or other >contamination. > >Ben > > >--- Benjamin Chui wrote: > > > Ed, > > > > The film actually consists of a Mo-AlN-Mo (molybdenum and > > aluminum > > nitride) stack. The tools that we expect to be using are: > > > > svgcoat > > EValign/KarlSuss > > svgdev > > > > drytek2 > > AMT > > wbmetal > > gasonics > > > > STSetch (any exposed metal will be fully encapsulated by photoresist, > > a > > method previously approved by Jim McVittie) > > > > Please advise if any of these material-tool combinations are a > > problem. > > > > Thanks! > > > > Ben > > > > > > --- Ed Myers wrote: > > > > > Ben, > > > > > > It's not the bringing them in to the fab, it's what tools they will > > > be > > > processed in and the level of cleanliness of the wafers. You will > > > need to > > > provide a process flow (tools) that will see the wafers. > > > > > > Regards, > > > > > > > > > At 12:01 PM 9/19/2005, you wrote: > > > >Hello specmat team, > > > > > > > > I'd like to request approval to bring wafers with deposited > > > aluminum > > > >nitride (AlN) film into the SNF clean room. I hope this is OK > > since > > > >users are specifically using AlN in place of PZT for its CMOS > > > >compatibility in building piezoelectric sensors. The company > > > >depositing the AlN also cites this advantage. > > > > > > > >Thanks, > > > > > > > >Ben Chui > > > > > > > > > > > > > > > > From yzhu at ee.ucr.edu Wed Sep 21 18:35:48 2005 From: yzhu at ee.ucr.edu (Yan Zhu) Date: Wed, 21 Sep 2005 17:35:48 -0800 Subject: Co silicide Message-ID: <20050922003550.C6BBD69F@post.ee.ucr.edu> Hi, I am a new member of the SNF. I will do some Cobalt silicide experiment, may I know if Cobalt silicide wafer can be loaded into the LTO furnace? And which machine is semi-clean, so I can deposit Cobalt before going to the LTO furnace? Thanks Yan Zhu yzhu at ee.ucr.edu From edmyers at stanford.edu Tue Sep 27 10:48:40 2005 From: edmyers at stanford.edu (Ed Myers) Date: Tue, 27 Sep 2005 10:48:40 -0700 Subject: SpecMat Log Sheet 9/27/05 Message-ID: <6.2.1.2.2.20050927103554.0425d4f0@edmyers.pobox.stanford.edu> Members, We have two new items for this week. Lets try and do this by email instead of holding todays meeting. Item 1) I'd like to request approval to bring wafers with deposited aluminum nitride (AlN) film into the SNF clean room. I hope this is OK since users are specifically using AlN in place of PZT for its CMOS compatibility in building piezoelectric sensors. The film actually consists of a Mo-AlN-Mo (molybdenum and aluminum nitride) stack. The tools that we expect to be using are: svgcoat, EValign/KarlSuss, svgdev, drytek2, AMT, wbmetal, gasonics and STSetch (any exposed metal will be fully encapsulated by photoresist, a method previously approved by Jim McVittie) Comments: Since Ben wants to run down the clean equipment set. This means he needs to prove the cleanliness of the deposition by TXRF. I will send out an email on Friday, Oct. 7th detailing the need for TXRF, or verification the film comes from an approved supplier if I don't hear any concerns. Item 2) Our process requires Cl-etching of Al stopping on Mg. Mg has shown that it does not etch in F-chemistry here and Cl-chemistry (Al,Ti, TiN, TaN etches) at ATDF. ATDF ran contamination checks and were satisfied that Mg did not pose a contamination risk to their mainline metal etcher. The chemistry I plan to use is BCl3 and Cl2 with a 40-80W RF power. In light of the contamination study, I would like to resubmit our request for Mg in the pquest. Comment: Carl has provided VPD-ICP-MS data for the wafers. The data meets our definition of Clean even though he wishes to run through the pQuest. I will give Jim a copy of the data. If I don't hear any reason we can't approve this request by Friday, Oct. 7th I will send out an approval email. Item 3) Is a carry over concerning cleaning pots for Ge. I will set a meeting with Mary and Nancy to come up with a strategy and propose it to SpecMat by Friday Oct. 7th. Regards, -------------- next part -------------- A non-text attachment was scrubbed... Name: SpecMat Logsheet.xls Type: application/octet-stream Size: 105984 bytes Desc: not available URL: From mdeal at stanford.edu Tue Sep 27 11:12:36 2005 From: mdeal at stanford.edu (Michael Deal) Date: Tue, 27 Sep 2005 11:12:36 -0700 Subject: SpecMat Log Sheet 9/27/05 In-Reply-To: <6.2.1.2.2.20050927103554.0425d4f0@edmyers.pobox.stanford.e du> References: <6.2.1.2.2.20050927103554.0425d4f0@edmyers.pobox.stanford.edu> Message-ID: <6.1.1.1.2.20050927111206.01e6c9f0@mdeal.pobox.stanford.edu> I have no problem with these conditions. -mike At 10:48 AM 9/27/2005, Ed Myers wrote: >Members, > >We have two new items for this week. Lets try and do this by email >instead of holding todays meeting. > >Item 1) I'd like to request approval to bring wafers with deposited >aluminum nitride (AlN) film into the SNF clean room. I hope this is OK >since users are specifically using AlN in place of PZT for its CMOS >compatibility in building piezoelectric sensors. The film actually >consists of a Mo-AlN-Mo (molybdenum and aluminum nitride) stack. The >tools that we expect to be using are: svgcoat, EValign/KarlSuss, svgdev, >drytek2, AMT, wbmetal, gasonics and STSetch (any exposed metal will be >fully encapsulated by photoresist, a method previously approved by Jim >McVittie) > >Comments: Since Ben wants to run down the clean equipment set. This means >he needs to prove the cleanliness of the deposition by TXRF. I will send >out an email on Friday, Oct. 7th detailing the need for TXRF, or >verification the film comes from an approved supplier if I don't hear any >concerns. > > >Item 2) Our process requires Cl-etching of Al stopping on Mg. Mg has >shown that it does not etch in F-chemistry here and Cl-chemistry (Al,Ti, >TiN, TaN etches) at ATDF. ATDF ran contamination checks and were >satisfied that Mg did not pose a contamination risk to their mainline >metal etcher. The chemistry I plan to use is BCl3 and Cl2 with a 40-80W >RF power. In light of the contamination study, I would like to resubmit >our request for Mg in the pquest. > >Comment: Carl has provided VPD-ICP-MS data for the wafers. The data >meets our definition of Clean even though he wishes to run through the >pQuest. I will give Jim a copy of the data. If I don't hear any reason >we can't approve this request by Friday, Oct. 7th I will send out an >approval email. > >Item 3) Is a carry over concerning cleaning pots for Ge. I will set a >meeting with Mary and Nancy to come up with a strategy and propose it to >SpecMat by Friday Oct. 7th. > >Regards, > > From jun.f.zheng at intel.com Fri Sep 30 10:47:21 2005 From: jun.f.zheng at intel.com (Zheng, Jun F) Date: Fri, 30 Sep 2005 10:47:21 -0700 Subject: Request for allowing poly and annealing process of wafers with CMOS compatible Si3N4 from outside Stanford CIS Message-ID: <88056F38E9E48644A0F562A38C64FB6005DECEE2@scsmsx403.amr.corp.intel.com> Sept. 30, 2005 To: Member of Special Materials Committee From: Jun-Fei Zheng Sub: Processing of Poly deposition on wafers received Si3N4 film from outside CIS We are evaluating Jet Vapor Deposited (JVD) Si3N4 film at Yale University for certain applications. The process is CMOS compatible in a custom built machine and wafers of 8" and 12" has been cross processed for CMOS application with several major semiconductor companies, including Intel. To illustared the details, the materials involved in the process are quartz, Teflon, and CMOS quality gas sources. JVD Si3N4 was process by mixture of SiH4 and Nitrogen gas in remote plasma through a quartz inner nozzle with He gas as carrying gas in an outer nozzle. The process is conducted at room temperature, which future reduce risk of high temperature diffusion induced contamination. We have tested the film contamination level by VPD (vapor phase decomposition) and all the metallic contents are below detecting limit or less than 10e^10/cm2, the criteria that we take such wafers into diffusion and poly furnace at Intel development line for further process without concerning the contamination based on the contamination test results. We only processed the wafer through SC1 and SC2 bath as precaution measure before inserted into Intel line. We are requesting to allow the same materials to be processed in CIS for in-situ poly deposition and annealing furnace. We want to run this at CIS because we need in-situ poly deposition capability. In order to fit the 6" tube or 4" tube, the 8" wafer has been notch cut and cleaned into half, with the short dimension of 4", so that can be inserted into 6" or 4" tube. We sincerely hope our request can be granted. We also wanted to avoid H2SO4:H2O2 process as our wafer has never been through Lithographic process and this is what the H2SO4:H2O2 mainly for but H2SO4:H2O2 process might destroy the gate quality Si3N4. We also do not see the need for 50:1 HF dip, which is used for clear oxide formed on Silicon during H2SO4:H2O2 process. We believe SC1 and SC2 in combination would be able to address potential contamination during wafer handling and shipping. If you have any suggestion and question, please reply to my e-mail or contact me at (408)921-8763 (cell). Sincerely, Jun-Fei Zheng Senior Staff Scientist Intel Corporation Strategic Technology Group SC1-05 3065 Bower Ave Santa Clara, CA 95052 -------------- next part -------------- An HTML attachment was scrubbed... URL: