Alex,

I talked with the staff and found that lower temperature depositions 
have been done in the past.  The low temperature does not appear to 
cause a problem with the equipment, but can have process problems, 
especially particles.  As a result, your group will be responsible to 
take all actions required to minimized potential particles for 
subsequent users.  This means, the chamber must be cleaned following 
your low temperature depositions (If particles are a problem for your 
process, you will need to lower the chamber temperature, let it 
stabilize, clean and coat the chamber before your deposition).

Following your low temperature deposition you must do the following 
before returning the system back to the user community.
1) do a chamber etch to remove any and all depositions on the chamber 
(the chamber temperature during the etch is not critical as long as 
all depositions have been removed)
2) Chamber coat per standard procedure.  The chamber must be at the 
standard temperature and equilibrated  before the chamber coatings 
can start and the films and film thickness must be per specification.

Regards,
Ed

At 06:41 PM 6/19/2007, Alexander Driskill-Smith wrote:
>Ed,
>
>Thanks for the brief discussion in the gowning room today. We're keen to try
>out the STS at 150 degrees C as soon as possible - I'm hoping sometime this
>week - so do please let me know as soon as you have spoken with the
>appropriate people. We can come in to the Fab anytime tomorrow (Wednesday),
>Thursday afternoon or anytime Friday to try out the process for the first
>time.
>
>Best regards,
>
>Alex
>
>
>The contents of this email message and any attachments are confidential and
>are intended solely for addressee. The information or ideas contained herein
>or attached hereto may also be legally privileged and will remain the
>property of the disclosing party. No license or other rights to any
>confidential information is granted or implied hereby. If you have received
>this transmission in error, any use, reproduction or dissemination of this
>transmission is strictly prohibited and enforceable by law. If you are not
>the intended recipient, please immediately notify the sender by reply email
>or at 408-945-2160 and delete this message and its attachments, if any.
>
>
>-----Original Message-----
>From: Alexander Driskill-Smith
>[mailto:alexander.driskill-smith at grandisinc.com]
>Sent: Monday, May 14, 2007 10:06 AM
>To: 'Ed Myers'
>Subject: RE: 150C PECVD
>
>Many thanks, Ed. This sounds promising. We'll keep an eye on Coral for
>updates.
>
>Best regards,
>
>Alex
>
>
>The contents of this email message and any attachments are confidential and
>are intended solely for addressee. The information or ideas contained herein
>or attached hereto may also be legally privileged and will remain the
>property of the disclosing party. No license or other rights to any
>confidential information is granted or implied hereby. If you have received
>this transmission in error, any use, reproduction or dissemination of this
>transmission is strictly prohibited and enforceable by law. If you are not
>the intended recipient, please immediately notify the sender by reply email
>or at 408-945-2160 and delete this message and its attachments, if any.
>
>
>-----Original Message-----
>From: Ed Myers [mailto:edmyers at stanford.edu]
>Sent: Monday, May 14, 2007 9:47 AM
>To: Alexander Driskill-Smith
>Subject: RE: 150C PECVD
>
>Alex,
>
>It looks like the system will be down for awhile.  We lost one of the RF
>generators and of course they are no longer supported.  One of Jeannie's
>concerns was making a change while the tool was unstable.  I had been voting
>the instability was this power supply.  When we get the power supply
>changed, I'm hoping we can eliminate Jeannie's concerns and you can move
>forward.
>
>Keep looking at Coral for updates on the STS status.
>
>Ed
>
>
>At 04:30 PM 5/1/2007, you wrote:
> >Ed,
> >
> >I have 3 quick questions for you.
> >
> >1) We will shortly need to order a couple of masks. I recall your
> >saying in the past that it's best to use an outside vendor for masks
> >and that it takes around 3-4 days at cost of around $750. Is that
> >correct? And which vendor would you recommend? On the SNF website, it
> >says that SNF users often go to Compugraphics, but I just wanted to
> >check that Compugraphics is still the preferred vendor.
> >
> >2) I'd like to bring our HSQ ebeam resist into SNF when we are next
> >visiting on Thursday (you approved it for SpecMat earlier this year).
> >But it needs to be stored in a refrigerator. Can you confirm there is a
> >refrigerator available and where it is located? I just want to check
> >this before we bring the bottle in.
> >
> >3) Following up on Yunfei's email on Friday, what is the procedure for
> >getting approval to run a 150 C recipe on the STS PECVD tool? I recall
> >that when Jeannie Perez trained us on this tool, she said something
> >about getting Review Board approval, but I'm not sure what the
> >procedure is.
> >
> >Best regards,
> >
> >Alex
> >
> >
> >
> >Alexander A. G. Driskill-Smith
> >Grandis, Inc.
> >1123 Cadillac Court, Milpitas, CA 95035
> >Telephone: (408) 945-2156
> >Facsimile: (408) 945-2161
> >Cellphone: (408) 807-4402
> >Email:
> ><blocked::mailto:alexander.driskill-smith at grandisinc.com>alexander.dris
> >kill-smith at grandisinc.com
> >Internet: <http://www.grandisinc.com/>www.grandisinc.com
> >
> >The contents of this email message and any attachments are confidential
> >and are intended solely for addressee. The information or ideas
> >contained herein or attached hereto may also be legally privileged and
> >will remain the property of the disclosing party. No license or other
> >rights to any confidential information is granted or implied hereby. If
> >you have received this transmission in error, any use, reproduction or
> >dissemination of this transmission is strictly prohibited and
> >enforceable by law. If you are not the intended recipient, please
> >immediately notify the sender by reply email or at 408-945-2160 and
> >delete this message and its attachments, if any.
> >
> >
> >-----Original Message-----
> >From: Yunfei Ding
> >[<mailto:yunfei.ding at grandisinc.com>mailto:yunfei.ding at grandisinc.com]
> >Sent: Friday, April 27, 2007 4:06 PM
> >To: emyers at snf.stanford.edu
> >Cc: 'Alex Driskill-Smith'; 'Shengyuan Wang'
> >Subject: 150C PECVD
> >
> >Hi Ed,
> >
> >Last year when we were discussing the feasibility of doing our process
> >in SNF we talked about the low temperature SiO2 deposition which is
> >critical for our process, you told us it's possible for us to try and
> >setup a 150C recipe on the STS PECVD tool. We are now working on the
> >process and have reached to this step. Please let me know what
> >precedure we need to go through for testing and setting up a new
> >recipe.
> >
> >We've done 150C PECVD on a plasmaquest tool, the recipe is listed here.
> >Please check if this is resonable or if you have any suggestion. If
> >possible we'd like to start with this recipe (or a recipe you
> >suggest) and make adjustments later on for better results.
> >
> >
> >Recipe:
> >gases:
> >         2% SiH4 in He 200 sccm
> >         N2O 450 sccm
> >pressure:
> >         900 mtorr
> >Power:
> >         20 watts
> >Tmeperaure:
> >         150C
> >
> >
> >
> >
> >
> >Best Regards,
> >
> >Yunfei Ding
> >
> >Grandis Inc.
> >1123 Cadillac Court
> >Milpitas, CA 95035
> >Phone: (408) 945 2158
> >Fax: (408) 945 2161
> >





More information about the specmat mailing list