Quest for processing in P5000
Mary Tang
mtang at stanford.edu
Thu Aug 4 21:02:03 PDT 2011
Hi all --
I would suggest that this is OK to process in p5000etch ChB because Pt
is not exposed to plasma and Pt is considered a "semiclean B" material
if it were deposited in a MOS-compatible backend system. If this was
deposited in innotec or metalica, it is technically "contaminated."
However, contamination of the etch chamber via sputtering under these
conditions is not very likely. That said, contamination of the PRS1000
bath when cleaning the resist in wbmetal could present risk
contamination, so I'd suggest for this case, that cleans be manually
done at wbgeneral (after matrix ashing.)
Any comments or concerns about this approach?
Mary
On 8/4/2011 7:37 PM, Lele Wang wrote:
> Dear Committee,
> I sent this email earlier but without reply. I tried other tools but it did not work so I am stuck on this step for 1 month.
>
> My wafer is in the contaminated group with Pt metal lines on top. I need to do a directional oxide etch at the bottom of the trench in P5000 and the wafer will be well protected by 1.6um photoresist on top. The opening area is a 30um deep, 5um wide silicon trench having thermal oxide on sidewall and bottom, which is clean. The bottom oxide to be etched is 800nm. Is that ok to etch this in p5000? Please reply ASAP. Thanks a lot.
>
>
>
> Best,
> Lele Wang
> Department of Electrical Engineering
More information about the specmat
mailing list