Test Results for Nitride, Low Freq. Oxide and High Freq. Oxide
Jeannie Perez
jperez at snf.stanford.edu
Mon Jan 7 15:47:48 PST 2008
Hi Everyone,
To my amazement after major PM changes to the STS system (new shower
head and realignment of Platen and Thermometer), the test results
were good. The GOOD news is that the Deposition Rate on all three
tests have increased, so take caution. I've never seen them so high.
Probably due to the alignment of the Platen. If you have concerns,
please notify me. I'll still be running more tests. Wafers measured on Woollam.
Low Frequency, four wafer averages;
Back Front Left side Right side Position on Platen
1798 1787 1791 1761 (Thickness)
1.47 1.47 1.47 1.47 (Index of Refraction)
0.847% 0.96% 0.84% 0.87% (range)
10 11,2 10 9.1 (std. dev.)
Low Freq. dep. rate is ~380 A / minute
Deposition Time was 4' 35"
High Frequency, four wafer averages;
Back Front Left side Right side Position on Platen
3864 3882 3918 3918 A (Thickness)
1.46 1.46 1.46 1.46 (Index of Refraction)
2.47% 1,40% 3,5% 3% (range)
57 30 78 72 (std Dev.)
High frequency dep. rate is ~390 A / minute
Deposition Time was 10 minutes.
Nitride Deposition for ~13.5 minutes- dep rate is ~ 126 to 130 A / minute
Only measured one of four tests on Woollam. Ed Myers will work on recipe.
Average thickness 1761A, Didn't give the correct Index of
Refraction. Ed figured out to be 1.97.
Regards,
Jeannie Perez
Science and Engineering Technician
Stanford Nanofabrication Facility
Stanford University
Tel: (650) 723-7997
Fax: (650) 725-6278