Comment stsetch SNF 2011-08-10 18:08:44: Completed monthly PM

eenriquez at snf.stanford.edu eenriquez at snf.stanford.edu
Wed Aug 10 18:08:45 PDT 2011


Found the helium pressure had been turned down to about 6 Torr instead of the maximum of 9.9 T.  When I adjusted the He pressure to the correct setting, the helium flow increased to 5.0.  This indicates a huge helium cooling leak.  When I opened the chamber, I found wafer chips underneath the wafer lifter.  This caused the lifter to sit slightly above the surface of the electrode.
Helium readings now are back to normal.  
2.3 : With the wafer loaded and before the process starts
3.7 - 4.0 :  During process (no wafer holder)
Nancy will run another test tomorrow