I worked on the SOI recipe yesterday. The success of the SOI recipe
seems to be VERY sensitive to your coil tuning parameters (not so
much with the Smooth-Shallow recipe). By changing the tuning value
by % moved me from a state where the etch plasma extinguished to one
which was stable. This tuning sensitivity was observed by STS while
they were on site. The copied our log files and carried them back to
the UK for analysis. I have not heard anything back from them.
You can use the SOI Template tuning points as a starting position if
your recipe is having problems. (By the way who deleted or renamed
the recipe) If your recipe has tuning problems you will see a yellow
fault indicating the coil could not reach set point, if you don't
stay around to watch the plasma. The downside is, it's the etch
cycle which is blinking off. Your etch rate will be drastically
reduced, while you are still getting your polymer formation. Be obserevant.
Ed