Fwd: Hotplates And Ovens Calibration
Eddy M Vilanova
vilanova at stanford.edu
Wed Jan 13 11:42:27 PST 2010
----- Forwarded Message -----
From: "Eddy M Vilanova" <vilanova at stanford.edu>
To: headway2 at snf.stanford.edu, Laurell-R at snf.stanford.edu, Coater at snf.stanford.edu
Cc: vilanova at snf.stanford.edu
Sent: Wednesday, January 13, 2010 11:41:02 AM GMT -08:00 US/Canada Pacific
Subject: Hotplates And Ovens Calibration
OVENS
Set Point 90oc Set Point After Calibration = 88oC
Set Point 110oC Set Point After Calibration = 108oC
Set Point 150oC Set POint After Calibration = 150.8oC
DEVELOPER TRACKS HOTPLATES
SVG DEV1 Setpoint 110oC Set Point After Calibration = 109oC
SVG DEV2 Setpoint 110oC Set POint After Calibration = 109.7oC
COAT TRACKS
90oC SetPOint after Calibration = 89.8oC
90oC Setpoint After Calibration = 90.1oC
DATA PLATES ACROSS FROM KALRSUSS ALIGNER
( FROM LEFT TO RIGHT )
Setpoint 110.oC ( Reading after calibration = 108.5)
Set POint 200oC ( Reading After Calibration = 205oC )
90oC Hotplate ( reading After Calibration = 91.5oC )
115oC Hotplate ( Reading After Calibration = 115.8oC )
HOTPLATES NEXT TO LAURELL SPINNER
200oC Setpoint ( reading after calibration = 198.0oC )
200oC Setpoint ( Reading After Calibration = 201oC )
More information about the svgcoat
mailing list