Problem tylan2 SNF 2009-10-15 14:37:33: MFS alarm : 08

seymour at snf.stanford.edu seymour at snf.stanford.edu
Thu Oct 15 14:48:41 PDT 2009


tube alarming for H2 temp. found user running wet700. Wioth tube idling  at 700 degrees I tried H2 flow and got alarm. Contacted Tystar and was advised to check H2 T/C voltage and found it to be 6.2 mV. Tystar advised to adjust R45 one turn clockwise and check operation. The MFS460 would then allow H2 to flow. 