For historical reasons I am adding the runs done earlier this and last
month. Please note, they are for WET900 00:42:00 and the measurements
were done on nanospec.
Folks,
Today I was able to run unifomity checks using Jim's recommended WET900
program. Both tubes were run with the same program for the same time
(42mins) at the same time. The results, below, are interesting.
The data is displayed with the front wafer first and the spread was over
two boats per tube, positioned identically;
Wf position Center Top Bottom Right Left Average
Tylan2
1 (front) 1061 1063 1058 1062 1061 1061A
2 1064 1066 1060 1067 1062 1064A
3 1072 1074 1069 1074 1070 1072A
4 (back) 1087 1091 1084 1090 1081 1087A
Tylan4
1 (front) 847 869 822 854 854 849A
2 871 884 854 879 876 873A
3 883 890 868 886 880 881A
4 (back) 889 893 883 889 891 889A
--
Nancy Latta