From rcrane at snf.stanford.edu Thu May 2 18:01:28 2002 From: rcrane at snf.stanford.edu (Dick Crane) Date: Thu, 02 May 2002 18:01:28 -0700 Subject: DCS flow problem Message-ID: <3CD1E168.35A048@snf.stanford.edu> Nitride users, At 0500 this morning and again around 1100 this morning the tylan nitride had a DCS MFC alarm and an out of spec flow rate. Both runs were of a long (6 hr) duration. A short (1 hr) dep was successfully run this afternoon. No obvious faults were found at the gas cylinder/purge panel or at the MFC. Gas pressure, cylinder and gas panel temperatures and product volume were all normal. Maintenance will run a six hour dep starting at 0630 tomorrow morning in an attempt to observe the failure mode and take corrective action. The nitride tube is not reliable at this time. If you wish to run tonight or until we find and repair the problem, do so at your own risk. Sorry for the inconvenience, Dick From rcrane at snf.stanford.edu Wed May 29 13:38:37 2002 From: rcrane at snf.stanford.edu (Dick Crane) Date: Wed, 29 May 2002 13:38:37 -0700 Subject: Current status DCS fault Message-ID: <3CF53C4D.2874AB1A@snf.stanford.edu> Nitride users: The DCS issue continues to be an annoyance. Last Sunday the DCS MFC faulted twice during one run in the afternoon. Given the time and temperature, DCS condensation was, most likely, not the cause. The maintenance crew has installed a datalogger on the DCS MFC to monitor control and response signal in hopes of catching and identifying the fault. As a user, you need not do anything differently. If you experience a DCS fault, please notify Ted or Ray. In the mean time the "Yellow" light remains on as the DCS problem may surface at anytime. Please use the nitride tube with this in mind. Thanks for your patience, Dick From tberg at snf.stanford.edu Thu May 30 08:44:50 2002 From: tberg at snf.stanford.edu (Ted Berg) Date: Thu, 30 May 2002 08:44:50 -0700 Subject: Nitride Tube Message-ID: <3CF648F2.89F6102B@snf.stanford.edu> Hello Users, The data logger is now connected to the DCS MFC. We successfully monitored the run last night and the flow exactly matched the setpoint throughout the run. We will continue to monitor all runs to further investigate flow issues. any issues please contact Ray, Dick , or myself. thanks,ted