Shutdown tylansige SNF 2008-10-04 18:48:02: Still error occurs

seymour at snf.stanford.edu seymour at snf.stanford.edu
Mon Oct 6 08:47:49 PDT 2008


found ch 4 N2 mfc to be above zero with no flow. Unable to adjust zero and it remain stable. replaced LM324 op amp and zero pot. adjusted zero and did auto zero routine. Putting system in pump recipe. 