From booth at snf.stanford.edu Tue Sep 2 10:19:12 2003 From: booth at snf.stanford.edu (booth at snf.stanford.edu) Date: Sep 2, 2003 10:19:12 AM Subject: Shutdown wafersaw 2003-08-29 15:42:46: Report Clear for wafersaw Message-ID: Flushed cooling water passages, flow checked OK. Installed new blade & system tested normally again. From glopez at snf.stanford.edu Fri Sep 5 12:09:50 2003 From: glopez at snf.stanford.edu (glopez at snf.stanford.edu) Date: Sep 5, 2003 12:09:50 AM Subject: Shutdown wafersaw 2003-09-05 00:09:49: Locked up Message-ID: System is frozen, occurred during height calibration. Pressed 'Stop' instead of 'Enter'. From booth at snf.stanford.edu Fri Sep 5 09:53:39 2003 From: booth at snf.stanford.edu (booth at snf.stanford.edu) Date: Sep 5, 2003 9:53:39 AM Subject: Shutdown wafersaw 2003-09-05 00:09:49: Report Clear for wafersaw Message-ID: Restarted system to clear fault. Checked & adjusted water jets slightly. From mtang at snf.stanford.edu Mon Sep 29 03:39:48 2003 From: mtang at snf.stanford.edu (mtang at snf.stanford.edu) Date: Sep 29, 2003 3:39:48 PM Subject: Shutdown wafersaw 2003-09-29 15:39:47: Blade is stuck Message-ID: Can't remove the blade. From cbaxter at snf.stanford.edu Mon Sep 29 08:45:10 2003 From: cbaxter at snf.stanford.edu (cbaxter at snf.stanford.edu) Date: Sep 29, 2003 8:45:10 PM Subject: Shutdown wafersaw 2003-09-29 15:39:47: Report Clear for wafersaw Message-ID: locking nut was over torc. loosen locking nut