From kimsangb at snf.stanford.edu Mon Jun 2 23:31:30 2008 From: kimsangb at snf.stanford.edu (kimsangb at snf.stanford.edu) Date: Mon, 2 Jun 2008 23:31:30 -0700 Subject: Problem woollam SNF 2008-06-02 23:31:29: calibration required Message-ID: I accidentally pressed power-off button when I tried to turn off the light. After that, if you try to run one of recipes, it would not proceed until it is calibrated. Sorry for the inconvenience. From maurice at snf.stanford.edu Wed Jun 4 14:04:11 2008 From: maurice at snf.stanford.edu (maurice at snf.stanford.edu) Date: Wed, 4 Jun 2008 14:04:11 -0700 Subject: Problem woollam SNF 2008-06-02 23:31:29: calibration required Message-ID: re cal'ed as per Ed's instructions From maurice at snf.stanford.edu Wed Jun 4 14:05:45 2008 From: maurice at snf.stanford.edu (maurice at snf.stanford.edu) Date: Wed, 4 Jun 2008 14:05:45 -0700 Subject: Comment woollam SNF 2008-06-04 14:05:41: re cal'ed as per Ed's instructions Message-ID: cal wafer measured 27.9 nm after system cal From emyers at snf.stanford.edu Mon Jun 23 11:24:31 2008 From: emyers at snf.stanford.edu (emyers at snf.stanford.edu) Date: Mon, 23 Jun 2008 11:24:31 -0700 Subject: Comment woollam SNF 2008-06-23 11:24:31: Lamp Changes 6/23/08 Message-ID: I changed the lamp at 968.9 hours of usage. Reset the lamp counter and calibrated system. I got a thickness of 27.1 nm verse an expected 27.4nm thickness. Let me know if you see any issues. From emyers at snf.stanford.edu Mon Jun 23 11:25:08 2008 From: emyers at snf.stanford.edu (emyers at snf.stanford.edu) Date: Mon, 23 Jun 2008 11:25:08 -0700 Subject: Comment woollam SNF 2008-06-04 14:05:41: re cal'ed as per Ed's instructions Message-ID: Noted. Lamp has been replaced since this calibration.