From edmyers at stanford.edu Mon Dec 3 14:40:58 2012 From: edmyers at stanford.edu (Ed Myers) Date: Mon, 03 Dec 2012 14:40:58 -0800 Subject: Woollam Insufficient Intensity Error Message-ID: <50BD2A7A.1030802@stanford.edu> All, Today I was resolving the Insufficient Intensity Error problem with the Woollam. There are a couple of points I want to reinforce at this time. 1) During the training it was emphasized to Never recalibrate the system. If you see this error it is most often happens when you are doing a lot of wafer mapping. There is a counter in the VASE Manager that calls for a calibration after a 100 or so measurements. If you encounter this request, simply close and reopen the VASE manager software. This resets the counter and you can continue on with the measurements. What happens if you do recalibrate the system? More than likely you have just adjusted the hardware/software and all further measurements are false. If you can not get out the recalibration request, contact a staff member. This is a much better approach than setting the system up to give all your lab mates bad measurements. 2) I could not determine the initial cause of the Insufficient Intensity Error. What I did find out, is someone made some mechanical adjustments on the system. This was not a simple turning of a knob by mistake, it was going and finding a wrench to make the adjustments. As with any tool within the SNF facility, this is against all rules and is not tolerated. We do have one lab member who leaves the Woollam with the system adjusted (software) for a different sample thickness. I will talk to this lab member to find out why this happens and to make sure they do not leave it in this state when they have completed their measurement. According to Coral history, this person did not use the Woollam during the latest event. I will also had the procedure (all software) in the comments section in Coral on how to recover from the error caused incorrect sample thickness. Regards, Ed