From emyers at snf.stanford.edu Wed Dec 10 15:30:19 2008 From: emyers at snf.stanford.edu (emyers at snf.stanford.edu) Date: Wed, 10 Dec 2008 15:30:19 -0800 Subject: Shutdown xactix SNF 2008-12-10 15:30:18: Bad Gauge Calibration Message-ID: A bad gauge calibration was ran, so the system does not know where it is at. It is so bad it can not get through the standard recalibration. I will have to go to the manual or call Xactix to see how to recover. From eenriquez at snf.stanford.edu Thu Dec 11 08:24:08 2008 From: eenriquez at snf.stanford.edu (eenriquez at snf.stanford.edu) Date: Thu, 11 Dec 2008 08:24:08 -0800 Subject: Shutdown xactix SNF 2008-12-10 15:30:18: Bad Gauge Calibration Message-ID: Ran an rough gauge calibration by setting the pump/purge cycle to 0 and the hard pump time to 10 min. Ran a standard calibration after the rough calibration routine completed.