Shutdown xactix SNF 2011-01-03 11:30:56: Will not initialize correctly

eenriquez at snf.stanford.edu eenriquez at snf.stanford.edu
Wed Jan 5 14:31:11 PST 2011


Found the valve manifold configuration file did not match the hardware.  Somehow the Product code and the OPEN Length values required by the device net enabled manifold had changed. I changed the product code from 416 to 332 and the Open Length from 2 to 4.  Calibrated gauges with no problems.
Will write up a troubleshooting procedure by 1/15 and place it in the Troubleshooting database.
Please check your recipes before processing your devices.